A Wooten
2008-08-07 14:10:12 UTC
I am receiving different measurements with my PXI-6115 (DAQ) device than I receive with my PXI-4070 (DMM).
I am attempting to measure a signal wich is supplied by the unit I am testing.
- The + side of the signal is input into the + input of the PXI-6115 or PXI-4070.
- The - side of the signal is tied to a common return (in the test equipment) which is then wired to the - input of the PXI-6115 or PXI-4070.
The PXI-6115 is measuring a value of 0.0 (which is what I expect to see)
The PXI-4070 is measuring a value of 0.60 which is out of range for the test. (I've also measured the signal with a hand held DMM and measured 0.6 volts)
Any Ideas why I am seeing these differences in measurements?
I've read some of the NI tutorials on how Differential and Pseudodifferential measurements minimize common mode voltages caused by ground loops. I understand that the PXI-6115 is a pseudodifferential device which is capable of rejecting common mode voltages. Isn't the PXI-4070 (DMM) also a differential measurement (which should also reject common mode voltages) ?
On a side note, I've also observed (using a desktop oscilloscope) that a signal's voltage will drop (on the scope) when I connect the signal to my PXI-6115. What's really interesting is that the size of the drop depends on how I've configured the input ranges of the PXI-6115. The voltage drops more when the PXI-6115 input range is set to +/-20v and less when the input range is set to +/-10v ( perhaps this is related the the issue described above).
Any help would be appreciated.
Thanks,
Anson
I am attempting to measure a signal wich is supplied by the unit I am testing.
- The + side of the signal is input into the + input of the PXI-6115 or PXI-4070.
- The - side of the signal is tied to a common return (in the test equipment) which is then wired to the - input of the PXI-6115 or PXI-4070.
The PXI-6115 is measuring a value of 0.0 (which is what I expect to see)
The PXI-4070 is measuring a value of 0.60 which is out of range for the test. (I've also measured the signal with a hand held DMM and measured 0.6 volts)
Any Ideas why I am seeing these differences in measurements?
I've read some of the NI tutorials on how Differential and Pseudodifferential measurements minimize common mode voltages caused by ground loops. I understand that the PXI-6115 is a pseudodifferential device which is capable of rejecting common mode voltages. Isn't the PXI-4070 (DMM) also a differential measurement (which should also reject common mode voltages) ?
On a side note, I've also observed (using a desktop oscilloscope) that a signal's voltage will drop (on the scope) when I connect the signal to my PXI-6115. What's really interesting is that the size of the drop depends on how I've configured the input ranges of the PXI-6115. The voltage drops more when the PXI-6115 input range is set to +/-20v and less when the input range is set to +/-10v ( perhaps this is related the the issue described above).
Any help would be appreciated.
Thanks,
Anson