One thing that helps me to make more accurate measurement on 9211 is to configure "Auto Zero Mode" on the "Device" tab. The default for "Auto Zero Mode" is "Once", that means it reads an internal auto zero channel for offset error at the beginning of the task and then subtract that offset from all reading from the module in that task. This is ideal for temperature that does change very much over time. Setting "Auto Zero Mode" to "Every Sample" will make the task to read the internal auto zero channel for offset error every sample and subtract the different offset error from each individual channel. However setting it to "Every Sample" adds an additional internal channel to the scanlist and that will lower the max sample rate for your task since 9211 is a scanning module. This is ideal for temperature that varies by a large amount over time. The offset error reading from the internal auto zero channel is usually very small at room temperature, but your testing environment has a more extreme temperature condition, the offset error might be fairly significant. You can check what the internal auto zero channel is reading by opening up a test panel in MAX and selected "9211/_aignd_vs_aignd" and looking at the acquired data. This returns data in volts. If the data return is the range of nanovolts, then the offset error is insignificant. If the data return is in the range of single digit microvolts, then the offset error is small enough that you can ignore depend on the requirement of your application. But if it is in tens of microvolts, then the offset error is fairly significant.